Quickly, glossing over details:
Input stages on DSOs (assuming they are the 1MOhm type, not 50 Ohm) are typically a resistive divider with an equivalen impedance of 1Mohm plus compensation capacitors on each leg to keep the bandpass level through the input frequency bandwidth (ie 100MHz, 200MHz,). The equivalent capacitance as seen looking "into" the scope should be around 11pf to 18pf.
The input resistors should be of high enough wattage so that they are not damaged by input voltages. The current flowing through a resistor will be bounded on one side by the input voltage and on the other by either ground or the internal protection clamp voltages. Like wise the capacitor should be able to withstand the voltage. Don't use wire wound resistors, low inductance metal film are pretty common. Need low temp. co resistors and capacitors.
Some scopes have multiple attenuators, switched by relays.
And the AC/DC coupling capacitor can be right at the input or in some cases after the attenuator.
There should be a pair of clamping voltages immedately after the input attenuator. The simplest clamps are done with diodes and zeners. Keep in mind that they can cause signal offset and also influence the node capacitance (variably I might add).
After the clamps there is typically a follower of some kind. In lower frequency designs it can be an op-amp buffer but higher frequency equipment often has a fet-follower pair. The FETs are bootstrapped to help improve offset accuracy but you sill may have some thermal drift.
Now you can begin your signal chain.
If you can get your hands on older Tektronix scope manuals they often have complete schematics. You can see the "state of the art" of 1980 in all its gorey detail. The concepts are still the same in more recent equipment but with improvements in bandwidth, speed, power dissipation, component size, etc.
Ebay is a good source for old manuals. Also a good source for used equipment in general.
Have fun.